Ultrasound
B-Scan giving 2D sectional image of foetus
Bright
area indicates large echo amplitude
Sound
waves are propagated at a speed of approximately 1540 m/sec in soft
tissues.
The
thickness, size and location of various soft tissue structures in relation
to the origin of the ultrasound beam are calculated at any point in
time using an 'echo technique'.
The
strength of the reflected sound wave depends on the difference in "acoustic
impedance" between adjacent structures. The acoustic impedance of a
tissue is related to its density; the greater the difference
in acoustic impedance between
two adjacent tissues the more reflective will be their boundary. Higher
frequencyultrasound waves have a longer near field and less divergence
in the far field; they permit better resolution
of small structures. More energy however is absorbed and scattered
by the soft tissues so that higher frequencies have less penetrating
ability.
Conversely, a transducer producing lower frequencies
will provide greater depth of penetration but less well defined images.
Focusing and aperture control technology are often employed to narrow
the beam along it's entire path to achieve maximum right-and-left (lateral)
resolution. The transducer of a real-time scanner typically contains
over 300 crystals( piezo electric devices) arranged in a row where each
emit and receives an ultrasound beam in rapid succession to form a sweep.
The part of the abdomen under the probe is "swept" about 30 times (frames)
a second and a moving picture (a real-time picture) will be formed (not
unlike the principle in a movie projector). Beam density and dynamic
range control technologies are further being incorporated into each
scanner's design to optimize the resultant image.
Ultrasound
is a non-invasive method (the patient does
not have to be cut open or have anything inserted into him/her) of monitoring
foetal development using ultrasonic acoustic waves. It does not employ
ionizing radiation, which would be dangerous foetus, as radiation risk
is greatest when much cell division occurs, and disruption of DNA would
be most harmful when organs are developing.
Ultrasound
is considered safe. The AIUM (American Institute of Ultrasound in Medicine)
in 1977 announced that "no independently confirmed significant biological
effects have been demonstrated in mammals" exposed to acoustic intensities
of less than 100 mW cm2.
Over the last twenty years collected data reveals that this is a conservative
upper limit.
Adverse
effects from ultrasound include:
localized
temperature rise in tissue,
cavitation
(growth of gas bubbles),
steady
radiation force exerted on tissue structures (which may result in
movement),
streaming
agitation in liquids,
shear
stress of objects in the streaming liquid and the
oscillatory
force of the sound field on all structures
but
these are only observed at acoustic intensities
well above the limit.
here
is no room for complacency, though. As with all scanning procedures exposure
times and intensities should be kept as low as possible. Patient exposure
should be minimized and logged carefully, especially in the case of pregnant
women. Output from machines can vary considerably and manufacturer's specifications
should be verified. The machine should be regularly checked to ensure
that output is 'true'.
Pregnancy
problems can be detected and general progress quantitatively monitored
by using A-scans for accurate measurements and B-scans for general development.
A
B-Scan produces a sectional 2-D image. Each point on the monitor represents
echo amplitude by 'grey-scale' representation. The brighter the point
on the screen the 'louder' the echo. This is used to identify the part
of the foetus to be measured. A multiple array of transducers is used to gain the information that builds up into the 2-d representation, whereas an A-scan probe only needs a single transducer..
A
simple ultrasound probe with a curved transducer array is ideal for
obstetric use as it has no moving parts, is comparatively inexpensive
and can be used by relatively unskilled operators. A frequency of between
1-3 MHz is suitable for abdominal imaging as low frequency produces
low resolution imaging and increasing attenuation occurs at very high
frequency.
Accurate
distances within the eye can be measured using a type of probe designed
especially for ophthalmic use. The scanner is either placed in direct
contact with the eye or via a water bath (less risk of damage to eye
surface) it therefore needs to be small and the transducer head suited
to the curvature of the eye. It is used in A-scan mode with a frequency
of 8-13MHz
The anatomical structure within the kidney can be
viewed using a common curvilinear (see below) probe with a frequency
of 5MHz in B-scan mode. The higher the frequency the better the resolution
of the image - 5 MHz will give detail of structure to within 1mm. A
simple scan technique is fine for this application.
The
function of the heart valve is best viewed using a trans-oesophageal
probe with a frequency of 2.0-5.0 MHz. M-Scan complemented by a B-scan.
Without this specialized probe a simple sector scan or compound scan
between the ribs would be necessary to avoid interference of them.
Doppler imaging would allow blood flow into and out of the valve to
be monitored and efficiency and any leakage can be assessed.
Doppler
Imaging uses the change in observed frequency of the ultrasound signal
to calculate the speed of flow:
The
angle q needs to be known if v is to be calculated
accurately. This is not easy and calculations of v are rarely performed.
Rather comparisons are made between v in comparable arteries and veins
to detect blockages.
The
factors that should be considered when selecting a frequency for Doppler
ultrasound imaging are:-
resolution
(better for larger frequency (about 1mm in soft tissue for 5MHz))
depth
of penetration (large frequency cannot probe as deeply).
Patient undergoing a kidney scan B-scan of liver and kidney
A
selection of ultrasound probes
Ultrasound
transducers, often called probes, come in different shapes and
sizes for use in different scanning situations. For example, in an obstetric
scan the probe used is usually one that looks like a curved soap bar
(more accurately known as a 'convex-array' transducer, and the one with
a flat surface a 'linear-array' transducer ) which can be slid over
the maternal abdomen while maintaining good surface contact to the abdominal
surface for the whole width of the probe. In a vaginal scan, the probe
has to be a long and slender piece to fit into the vagina. Pulsed ultrasound,
because of it's high frequency can be aimed in a specific direction
and obeys the laws of geometric optics with regard to reflection, transmission
and refraction. When an ultrasound wave meets an interface of 'differing
echogenicity' (big difference in acoustic impedance between the media),
the wave is reflected, refracted or absorbed. Reflected sound waves
are processed. The transducer, though emitting ultrasound in rapid pulses,
acts as a receiver most of the time. The ultrasound images can be displayed
on an oscilloscope screen or a video monitor (via what is known as a
scan converter) and can be recorded on videotape, thermal paper or radiographic
film.
B
and M mode images of the heart at the level of the mitral valve
M-mode image of the heart
A
trans-oesophageal probe - so that the transducer array is nearer to
the patient's heart
the narrow flexible tube is swallowed by the patient. This avoids
having to
scan between the gaps in a patients ribs.
NB:
M-mode is not on the syllabus neither is the Doppler equation - they
are in these notes to extend your background knowledge. Calculations
will be restricted to simple echo or simple Doppler equation applications.
Useful
Websites to visit:
http://www.ob-ultrasound.net/
The
following is an extract from this site (taken in March 1999). The site
is well worth a visit!
Doppler
Ultrasound
The
doppler shift principle has been used for a long time in foetal heart
rate detectors. Further developments in doppler ultrasound technology
in recent years have enabled a great expansion in it's application in
Obstetrics, this time in the area of assessing and monitoring the well
being of the foetus.
Blood flow characteristics in the foetal blood vessels can be assessed
with Doppler 'flow velocity waveforms'. Diminished flow, particularly
in the diastolic phase of a pulse cycle is associated with compromise
in the foetus. Various ratios of the systolic to diastolic flow are
used as a measure of this
compromise. The blood vessels commonly interrogated include the umbilical
artery, the aorta, the middle cerebral arteries and the uterine arcuate
arteries.
The use of colour flow mapping can clearly depict the flow of blood
in foetal blood vessels in a real-time scan, the direction of the flow
being represented by different colours. 'Colour' doppler is particularly
indispensable in the diagnosis and assessment of congenital heart abnormalities.
Another recent development is the Power Doppler (Doppler angiography).
It uses amplitude information from doppler signals rather than flow
velocity information to visualize slow flow in smaller vessels. A colour
perfusion-like display of a particular organ such as the placenta overlapping
on the 2-D image can be very nicely depicted.
Doppler
blood flow velocity waveforms
The
integration of real-time imaging with pulsed wave Doppler allows for
the vessel of interest to be identified and as the diameter of the vessel
and the angle of interrogation can be measured, it allows for an estimation
of velocity and volume of flow. This, nevertheless is not done in Obstetric
scans because of the difficulty and inconsistency encountered in non-linear
blood vessels such as the umbilical artery. As the foetus is completely
dependent on the supply of oxygen and nutrients from the placenta, examination
of the blood flow through the umbilical circulation would appear to
have great potential for the assessment of foetal health.
In
the umbilical artery, there is a relatively high forward velocity during
diastole, consistent with blood flow into a low-impedance vascular bed,
the placenta. With advancing gestation, there is an increase in end-diastolic
flow velocity relative to peak systolic velocity. This is attributed
to decreased resistance in the placental circulation with advancing
gestation. This change in the pulse velocity waveform can be quantified
by the systolic-to-diastolic ratio (A/B ratio). In pregnancies in which
the A/B ratio is elevated, there is an increase in intrauterine growth
compromise due to a placental circulation that has diminished in volume
owing to placental vascular occlusion.
What
about Safety?
It
has been over 35 years since ultrasound was first used on pregnant women.
Unlike X-rays, ionizing irradiation is not present and embryotoxic effects
associated with such irradiation should not be relevant. The use of
high intensity ultrasound is associated with the effects of "cavitation"
and
"heating" which can be present with prolonged insonation in laboratory
situations. Harmful effects in cells of experimental animals or humans
however have not been demonstrated in the large amount of studies that
have so far appeared in the medical literature purporting to the use
of diagnostic
ultrasound in the clinical setting. Findings in one study reporting
lower birth-weights in babies exposed to prenatal ultrasound have not
been reproducible.
Nevertheless it is general consensus that ultrasound scans should best
be performed when there is an indication to do so.
Colour
imaging
This
is a recent addition to plain 2-D real-time scanning. Also known as
"chroma" scans, user-selectable colour hues are assigned to the shades
of grey for better visualization of subtle tissue details. This clever
enhancement is aimed at better interpretation of the scans. 'Colour
scans' do not imply that various
parts of the same picture are depicted in different colours like what
we see in a colour photograph.
3-D
Ultrasound
3-d Image of the spinal column
3-dimensional
ultrasound is slowly moving out of the research and development stages
and is very much in the News. Faster and more advanced commercial models
are coming into the market. The scans requires special probes and software
to accumulate and render the images, and the rendering
time has been reduced from minutes to seconds. A good 3D image is often
quite impressive and further 2D scans may be extracted from 3D blocks
of scanned information. Volumetric measurements are more accurate and
both doctors and parents can better appreciate a certain abnormality
or the
absence of a certain abnormality in a 3D scan than a 2D one and there
is the possibility of increasing psychological bonding between the parents
and the baby. A large volume of literature and documentation is expected
to come out in the coming years and the diagnosis of congenital anomalies
could receive revived attention. Present evidence has already suggested
that even small defects such as spina bifida, cleft lips/palate, and
polydactyl may be more lucidly demonstrated. Other more subtle features
such as low-set ears, facial dysmorphia or clubbing of feet can be better
assessed, leading to
more effective diagnosis of chromosomal abnormalities. The study of
foetal cardiac malformations is also receiving attention. The ability
to obtain a good 3D picture is nevertheless still very much dependent
on operator skill, the amount of liquor around the foetus, it's position
and the degree of maternal obesity, so that a good image is not always
readily obtainable.
Other
experts in this field have not considered that 3D ultrasound will be
a mandatory evolution of our conventional 2D scans, rather it is an
additional piece of tool like doppler ultrasound. Whether 3D ultrasound
will provide unique information or merely supplemental information will
only wait to be seen. It's greatest potential is still in research and
particularly in the study of foetal embryology.